I decided it was time to recalibrate my DP4 probe (a DC3IO system I've had for over 10 years). I checked it against the ring gauge a few days ago and it was .0005" off, and it used to be much closer than that.

The manual instructions say to disconnect the cable from the probe.

It has been many years since I've done this and I can not remember how the FFG.0B.305.CLAD56 connector is supposed to release. I thought it might be a bayonet fitting (Digi-Key says it is a "push-pull" but I'm getting nowhere with that, and attempting to unscrew the barrel has it, the red dot and the cable spin so if it is a push-pull the little nubs may have been sheared off. On top of that, now when I ALT-I I don't see any red/green light changes on the inputs, the light on the probe doesn't come on, and the machine can't see that it is attached. I fear I've damaged the plug.

If someone can tell me how this thing (correctly) comes apart maybe then I can tell if I've damaged the probe.

So much for my good intentions of "I'll get the probe freshly calibrated for the start of the new year". :(

thanks,
Michael

Re: DP4 connector help

Posted: Sun Dec 27, 2020 12:55 am
by cncsnw
It pulls straight out. There is no locking mechanism; just a ramp that forms a slight detent.

Re: DP4 connector help

Posted: Sun Dec 27, 2020 1:35 am
by Michael Moore
Should I pull with the two red dots aligned? I'm pulling as hard as I can (bare hands, no pliers, stylus removed for safety) and it doesn't budge, whether the dots are aligned or not. Wiggling the connector while pulling on it doesn't help. Is the connector supposed to be able to spin freely like an RCA stereo type of jack, because I'm not detecting any binding when it is turned. The outer barrel with the 4 grooves in it has about .01-.02" of axial play in it, if that indicates anything.

Stuck seems to be a good description.

thanks,
Michael

Re: DP4 connector help

Posted: Sun Dec 27, 2020 9:28 am
by tblough
Normally, the red dots remain aligned and neither turns. You remove the plug by pulling slightly on the outer barrel of the plug (that's the 0.020 axial "play"). The fact that the outer barrel now rotates indicates something is seriously wrong with the connector.

I suspect the wiring within either the plug or jack (possibly both) is now twisted/broken and is shorting. The AllIn1DC has a fuse on the probe power circuit, but I don't know about the DC3IO. Either/both would account for the lack of I/O changes and LED activity on the probe.

I would suggest sending your probe and cable back to Centroid for repair.

Re: DP4 connector help

Posted: Sun Dec 27, 2020 12:45 pm
by Michael Moore
Hi Tom,

That was the conclusion I'd come to. At least I don't have anything pending where I will be held up while the probe is out for service.